Could anyone shed some light on this please.
Protrack keep failing on the same voltage and some times on the 2nd voltage after its been on for longer.
And if it's been on for longer tge voltage it fails at is a bit higher.
Please see picture of failed voltages.
Those voltages are reading just a bit high. You could try running through the ProTrack setup procedure:
ProTrack final test procedure (https://huntron.com/privatesales/technical/protrack-final-test-procedure.pdf)
You would be looking at section 2.0.
Good morning, thank you for your reply.
Looks like I got to do that on a day off. ;D ;D
Happy holidays
Thank you so much for the information.
preformed an Auto Cal it fixed the voltage check failing.
Thank you fetron for sharing that one in previus posts a while back.
Later, I will try to find and put link to that post here.
https://huntron.com/forum/index.php?topic=324.msg1339#msg1339