Has anyone using Workstation 4.3 on a ProTrack + Scanner fixed the Sig Check failure on the Hardware Check.
Huntron Tech tell me this is not a problem to "lose sleep over", but just curious to know if anyone has fixed the problem and how they did it.

admin

#1
While I am not too worried about it, I am still looking to see if I can get Sig Check to pass on my ProTrack model 20 (KJ2 vintage). I have 5 high resistance, high frequency steps that fail by about 1-4 points (MaxDiff is 6 to 10 with Tol=6). The range combinations suggest the issue is split trace (looped signature) that is more prevalent in high resistance, high frequency range combinations. To view your failed steps, click the View Log button after the Hardware Check has been run. Scroll to the bottom to see that most recent tests. I would be interested to see how my results compare to others. Here are the failed steps:

          V=13 R=11 F=38 Channel=B, MaxDif=7 Tol=6 Failed ***
          V=23 R=12 F=38 Channel=B, MaxDif=1 Tol=6 Passed
          V=0 R=0 F=39 Channel=A, MaxDif=3 Tol=6 Passed
          V=5 R=10 F=39 Channel=A, MaxDif=10 Tol=6 Failed ***
          V=13 R=11 F=39 Channel=A, MaxDif=8 Tol=6 Failed ***
          V=23 R=12 F=39 Channel=A, MaxDif=1 Tol=6 Passed
          V=0 R=0 F=39 Channel=B, MaxDif=2 Tol=6 Passed
          V=5 R=10 F=39 Channel=B, MaxDif=10 Tol=6 Failed ***
          V=13 R=11 F=39 Channel=B, MaxDif=9 Tol=6 Failed ***
The most important troubleshooting tool is your eyes.

jvthorsley

#2
Hi Curtis

I have copied here the text file for the Sig Check on my ProTrack + Scanner using Workstation 4.3.

          *** SIG CHECK ***
          V=0 R=0 F=0 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=0 Channel=A, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=0 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=0 Channel=A, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=0 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=0 Channel=B, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=0 Channel=B, MaxDif=3 Tol=6 Passed
          V=23 R=12 F=0 Channel=B, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=1 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=1 Channel=A, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=1 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=1 Channel=A, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=1 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=1 Channel=B, MaxDif=3 Tol=6 Passed
          V=13 R=11 F=1 Channel=B, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=1 Channel=B, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=2 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=2 Channel=A, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=2 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=2 Channel=A, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=2 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=2 Channel=B, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=2 Channel=B, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=2 Channel=B, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=3 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=3 Channel=A, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=3 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=3 Channel=A, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=3 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=3 Channel=B, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=3 Channel=B, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=3 Channel=B, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=4 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=4 Channel=A, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=4 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=4 Channel=A, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=4 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=4 Channel=B, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=4 Channel=B, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=4 Channel=B, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=5 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=5 Channel=A, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=5 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=5 Channel=A, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=5 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=5 Channel=B, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=5 Channel=B, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=5 Channel=B, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=6 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=6 Channel=A, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=6 Channel=A, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=6 Channel=A, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=6 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=6 Channel=B, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=6 Channel=B, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=6 Channel=B, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=7 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=7 Channel=A, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=7 Channel=A, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=7 Channel=A, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=7 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=7 Channel=B, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=7 Channel=B, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=7 Channel=B, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=8 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=8 Channel=A, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=8 Channel=A, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=8 Channel=A, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=8 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=8 Channel=B, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=8 Channel=B, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=8 Channel=B, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=9 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=9 Channel=A, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=9 Channel=A, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=9 Channel=A, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=9 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=9 Channel=B, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=9 Channel=B, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=9 Channel=B, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=10 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=10 Channel=A, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=10 Channel=A, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=10 Channel=A, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=10 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=10 Channel=B, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=10 Channel=B, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=10 Channel=B, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=11 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=11 Channel=A, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=11 Channel=A, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=11 Channel=A, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=11 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=11 Channel=B, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=11 Channel=B, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=11 Channel=B, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=12 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=12 Channel=A, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=12 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=12 Channel=A, MaxDif=6 Tol=6 Passed
          V=0 R=0 F=12 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=12 Channel=B, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=12 Channel=B, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=12 Channel=B, MaxDif=6 Tol=6 Passed
          V=0 R=0 F=13 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=13 Channel=A, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=13 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=13 Channel=A, MaxDif=6 Tol=6 Passed
          V=0 R=0 F=13 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=13 Channel=B, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=13 Channel=B, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=13 Channel=B, MaxDif=6 Tol=6 Passed
          V=0 R=0 F=14 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=14 Channel=A, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=14 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=14 Channel=A, MaxDif=6 Tol=6 Passed
          V=0 R=0 F=14 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=14 Channel=B, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=14 Channel=B, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=14 Channel=B, MaxDif=6 Tol=6 Passed
          V=0 R=0 F=15 Channel=A, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=15 Channel=A, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=15 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=15 Channel=A, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=15 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=15 Channel=B, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=15 Channel=B, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=15 Channel=B, MaxDif=6 Tol=6 Passed
          V=0 R=0 F=16 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=16 Channel=A, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=16 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=16 Channel=A, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=16 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=16 Channel=B, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=16 Channel=B, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=16 Channel=B, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=17 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=17 Channel=A, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=17 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=17 Channel=A, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=17 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=17 Channel=B, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=17 Channel=B, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=17 Channel=B, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=18 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=18 Channel=A, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=18 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=18 Channel=A, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=18 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=18 Channel=B, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=18 Channel=B, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=18 Channel=B, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=19 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=19 Channel=A, MaxDif=5 Tol=6 Passed
          V=13 R=11 F=19 Channel=A, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=19 Channel=A, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=19 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=19 Channel=B, MaxDif=5 Tol=6 Passed
          V=13 R=11 F=19 Channel=B, MaxDif=6 Tol=6 Passed
          V=23 R=12 F=19 Channel=B, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=20 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=20 Channel=A, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=20 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=20 Channel=A, MaxDif=6 Tol=6 Passed
          V=0 R=0 F=20 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=20 Channel=B, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=20 Channel=B, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=20 Channel=B, MaxDif=7 Tol=6 Failed ***
          V=0 R=0 F=21 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=21 Channel=A, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=21 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=21 Channel=A, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=21 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=21 Channel=B, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=21 Channel=B, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=21 Channel=B, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=22 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=22 Channel=A, MaxDif=6 Tol=6 Passed
          V=13 R=11 F=22 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=22 Channel=A, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=22 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=22 Channel=B, MaxDif=7 Tol=6 Failed ***
          V=13 R=11 F=22 Channel=B, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=22 Channel=B, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=23 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=23 Channel=A, MaxDif=6 Tol=6 Passed
          V=13 R=11 F=23 Channel=A, MaxDif=3 Tol=6 Passed
          V=23 R=12 F=23 Channel=A, MaxDif=6 Tol=6 Passed
          V=0 R=0 F=23 Channel=B, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=23 Channel=B, MaxDif=6 Tol=6 Passed
          V=13 R=11 F=23 Channel=B, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=23 Channel=B, MaxDif=6 Tol=6 Passed
          V=0 R=0 F=24 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=24 Channel=A, MaxDif=5 Tol=6 Passed
          V=13 R=11 F=24 Channel=A, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=24 Channel=A, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=24 Channel=B, MaxDif=4 Tol=6 Passed
          V=5 R=10 F=24 Channel=B, MaxDif=6 Tol=6 Passed
          V=13 R=11 F=24 Channel=B, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=24 Channel=B, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=25 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=25 Channel=A, MaxDif=5 Tol=6 Passed
          V=13 R=11 F=25 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=25 Channel=A, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=25 Channel=B, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=25 Channel=B, MaxDif=5 Tol=6 Passed
          V=13 R=11 F=25 Channel=B, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=25 Channel=B, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=26 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=26 Channel=A, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=26 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=26 Channel=A, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=26 Channel=B, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=26 Channel=B, MaxDif=5 Tol=6 Passed
          V=13 R=11 F=26 Channel=B, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=26 Channel=B, MaxDif=6 Tol=6 Passed
          V=0 R=0 F=27 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=27 Channel=A, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=27 Channel=A, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=27 Channel=A, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=27 Channel=B, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=27 Channel=B, MaxDif=4 Tol=6 Passed
          V=13 R=11 F=27 Channel=B, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=27 Channel=B, MaxDif=3 Tol=6 Passed
          V=0 R=0 F=28 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=28 Channel=A, MaxDif=6 Tol=6 Passed
          V=13 R=11 F=28 Channel=A, MaxDif=3 Tol=6 Passed
          V=23 R=12 F=28 Channel=A, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=28 Channel=B, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=28 Channel=B, MaxDif=7 Tol=6 Failed ***
          V=13 R=11 F=28 Channel=B, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=28 Channel=B, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=29 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=29 Channel=A, MaxDif=6 Tol=6 Passed
          V=13 R=11 F=29 Channel=A, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=29 Channel=A, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=29 Channel=B, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=29 Channel=B, MaxDif=6 Tol=6 Passed
          V=13 R=11 F=29 Channel=B, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=29 Channel=B, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=30 Channel=A, MaxDif=6 Tol=6 Passed
          V=5 R=10 F=30 Channel=A, MaxDif=5 Tol=6 Passed
          V=13 R=11 F=30 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=30 Channel=A, MaxDif=6 Tol=6 Passed
          V=0 R=0 F=30 Channel=B, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=30 Channel=B, MaxDif=6 Tol=6 Passed
          V=13 R=11 F=30 Channel=B, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=30 Channel=B, MaxDif=6 Tol=6 Passed
          V=0 R=0 F=31 Channel=A, MaxDif=6 Tol=6 Passed
          V=5 R=10 F=31 Channel=A, MaxDif=8 Tol=6 Failed ***
          V=13 R=11 F=31 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=31 Channel=A, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=31 Channel=B, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=31 Channel=B, MaxDif=8 Tol=6 Failed ***
          V=13 R=11 F=31 Channel=B, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=31 Channel=B, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=32 Channel=A, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=32 Channel=A, MaxDif=7 Tol=6 Failed ***
          V=13 R=11 F=32 Channel=A, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=32 Channel=A, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=32 Channel=B, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=32 Channel=B, MaxDif=8 Tol=6 Failed ***
          V=13 R=11 F=32 Channel=B, MaxDif=6 Tol=6 Passed
          V=23 R=12 F=32 Channel=B, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=33 Channel=A, MaxDif=6 Tol=6 Passed
          V=5 R=10 F=33 Channel=A, MaxDif=7 Tol=6 Failed ***
          V=13 R=11 F=33 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=33 Channel=A, MaxDif=6 Tol=6 Passed
          V=0 R=0 F=33 Channel=B, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=33 Channel=B, MaxDif=7 Tol=6 Failed ***
          V=13 R=11 F=33 Channel=B, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=33 Channel=B, MaxDif=6 Tol=6 Passed
          V=0 R=0 F=34 Channel=A, MaxDif=6 Tol=6 Passed
          V=5 R=10 F=34 Channel=A, MaxDif=7 Tol=6 Failed ***
          V=13 R=11 F=34 Channel=A, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=34 Channel=A, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=34 Channel=B, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=34 Channel=B, MaxDif=7 Tol=6 Failed ***
          V=13 R=11 F=34 Channel=B, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=34 Channel=B, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=35 Channel=A, MaxDif=6 Tol=6 Passed
          V=5 R=10 F=35 Channel=A, MaxDif=6 Tol=6 Passed
          V=13 R=11 F=35 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=35 Channel=A, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=35 Channel=B, MaxDif=6 Tol=6 Passed
          V=5 R=10 F=35 Channel=B, MaxDif=6 Tol=6 Passed
          V=13 R=11 F=35 Channel=B, MaxDif=3 Tol=6 Passed
          V=23 R=12 F=35 Channel=B, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=36 Channel=A, MaxDif=6 Tol=6 Passed
          V=5 R=10 F=36 Channel=A, MaxDif=5 Tol=6 Passed
          V=13 R=11 F=36 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=36 Channel=A, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=36 Channel=B, MaxDif=5 Tol=6 Passed
          V=5 R=10 F=36 Channel=B, MaxDif=6 Tol=6 Passed
          V=13 R=11 F=36 Channel=B, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=36 Channel=B, MaxDif=4 Tol=6 Passed
          V=0 R=0 F=37 Channel=A, MaxDif=7 Tol=6 Failed ***
          V=5 R=10 F=37 Channel=A, MaxDif=6 Tol=6 Passed
          V=13 R=11 F=37 Channel=A, MaxDif=4 Tol=6 Passed
          V=23 R=12 F=37 Channel=A, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=37 Channel=B, MaxDif=6 Tol=6 Passed
          V=5 R=10 F=37 Channel=B, MaxDif=6 Tol=6 Passed
          V=13 R=11 F=37 Channel=B, MaxDif=5 Tol=6 Passed
          V=23 R=12 F=37 Channel=B, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=38 Channel=A, MaxDif=6 Tol=6 Passed
          V=5 R=10 F=38 Channel=A, MaxDif=10 Tol=6 Failed ***
          V=13 R=11 F=38 Channel=A, MaxDif=6 Tol=6 Passed
          V=23 R=12 F=38 Channel=A, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=38 Channel=B, MaxDif=6 Tol=6 Passed
          V=5 R=10 F=38 Channel=B, MaxDif=11 Tol=6 Failed ***
          V=13 R=11 F=38 Channel=B, MaxDif=7 Tol=6 Failed ***
          V=23 R=12 F=38 Channel=B, MaxDif=6 Tol=6 Passed
          V=0 R=0 F=39 Channel=A, MaxDif=8 Tol=6 Failed ***
          V=5 R=10 F=39 Channel=A, MaxDif=11 Tol=6 Failed ***
          V=13 R=11 F=39 Channel=A, MaxDif=8 Tol=6 Failed ***
          V=23 R=12 F=39 Channel=A, MaxDif=5 Tol=6 Passed
          V=0 R=0 F=39 Channel=B, MaxDif=6 Tol=6 Passed
          V=5 R=10 F=39 Channel=B, MaxDif=12 Tol=6 Failed ***
          V=13 R=11 F=39 Channel=B, MaxDif=8 Tol=6 Failed ***
          V=23 R=12 F=39 Channel=B, MaxDif=5 Tol=6 Passed
          Sig Check FAILED ***
          Data Check Passed
          Phase Check Passed
          Res Check Passed
          Volts Check Passed
          Hz Check Passed
          Pulse Check Passed
          Scanner Test Passed
          Hardware Check Pass #1 FAILED *** at 17:35:48 on 20/12/2019

As you say its not essential to pass this test, just irritating that I have not found a way to fix it.

          ------------------------------
         

admin

My Hardware Checks were run without the Scanner connected as it checks the ProTrack only.
The most important troubleshooting tool is your eyes.

jvthorsley

Hi Curtis
If it helps get to the source of the problem I will run the test again without the Scanner connected.

jvthorsley

#5
Hi Curtis
I have run the ProTrack hardware tests this time without my Scanner attached.
And put on attachment here the text file showing the results of the failed SIG TEST.
Ran this test using Workstation Ver. 4.3.7278.13947.