Huntron User Support Forums => Out of Production Hardware => Topic started by: Al1234 on November 18, 2023, 11:28:42 AM

Title: ProTrack I, hi Freq, hi Resis
Post by: Al1234 on November 18, 2023, 11:28:42 AM
Hi every one. The protrack kj4 specialy shows capacitance in both channels in higher Frequency and resistance setting.
The kj3 unit is not as bad, as seen in the pictures.
Any suggestion where and how to start correcting this issue.
There is a tiny difference when leads are connected than when not. Seen in pictures.
Title: Re: ProTrack I, hi Freq, hi Resis
Post by: fetron on November 19, 2023, 12:23:53 AM
Hi!

Please warm up 20 min.
Next step: run the Autocal and testing.
Title: Re: ProTrack I, hi Freq, hi Resis
Post by: jvthorsley on November 19, 2023, 01:59:33 PM
Sounds a good idea from Fetron.
Can't wait to hear the next episode!!
Title: Re: ProTrack I, hi Freq, hi Resis
Post by: admin on November 20, 2023, 06:38:42 AM
Needs the Loop Calc to be run from Workstation and when prompted if you want to upload the calc settings to the ProTrack click yes (or OK, doing this from memory).
Title: Re: ProTrack I, hi Freq, hi Resis
Post by: Al1234 on November 20, 2023, 12:13:40 PM
Thank you. I have done this before I think. But I will do it again and report back once done.
Title: Re: ProTrack I, hi Freq, hi Resis
Post by: Al1234 on November 20, 2023, 01:04:32 PM
Perfect that fixed it. Thank you all. Though when I set it up on 100Kohms the graph shows just a slight resistance. And back to 50Kohms line stays flat.
But It's a 1000% better now.
Title: Re: ProTrack I, hi Freq, hi Resis
Post by: admin on November 20, 2023, 02:17:13 PM
The slight angle at 100Kohms at 5KHz is pretty much normal as it was difficult to compensate for low level capacitance on an analog CRT. With the current digital Trackers it is not an issue.
Title: Re: ProTrack I, hi Freq, hi Resis
Post by: Al1234 on November 20, 2023, 03:17:04 PM
Thank you
Title: Re: ProTrack I, hi Freq, hi Resis
Post by: Al1234 on November 30, 2023, 11:10:26 AM
Good morning, I have another question regarding the ProTrack (KJ4 or KJ3), setting. Perhaps it needs it's own topic.
The factory default setup has resistance:
Low    54 Ohm
Mid 1  1.24 KOhm
Mid 2  27.6 KOhm

These resistance setting are not selectable through the encoder nub.
Can users set up resistance besides the 13 selectable ones?
Title: Re: ProTrack I, hi Freq, hi Resis
Post by: admin on November 30, 2023, 02:47:47 PM
Yes. There are 20 Groups beside SMT, Passive, Factory, TLL, Caps, etc. that can be configured by the user. This is done through the front panel. Press Menu, then Group Edit, then select the Group (Button just below the Menu button) and rotate the encoder until you see Group 20. You can edit this one or any of the other 19 Groups. Use the Name button to change the Group name if you want.
Title: Re: ProTrack I, hi Freq, hi Resis
Post by: admin on February 29, 2024, 09:11:11 AM
To elaborate a bit further on this topic, you can set up your own Groups but you can also set up your own custom Ranges. Press the MENU button, then Range Edit. You have 100 spots for custom ranges. Select the Range position in the upper left, the settings on the right side and change the name on the bottom left. These custom ranges can be added to a custom Group.

This is some of the cool stuff the ProTrack had that was not carried over to newer products.
Title: Re: ProTrack I, hi Freq, hi Resis
Post by: jvthorsley on March 01, 2024, 07:19:48 AM
Hi Curtis
Your comment "This is some of the cool stuff the ProTrack had that was not carried over to newer products".
Reminds me you can't take the output from the internal Scanner on the 3200S to a DMM or counter.
Perhaps when the warranty period has expired someone might have a look inside the unit and see if it can be done as on the ProTrack Scanner.
Title: Re: ProTrack I, hi Freq, hi Resis
Post by: admin on March 04, 2024, 06:27:15 AM
Quote from: jvthorsley on March 01, 2024, 07:19:48 AMHi Curtis
Your comment "This is some of the cool stuff the ProTrack had that was not carried over to newer products".
Reminds me you can't take the output from the internal Scanner on the 3200S to a DMM or counter.
Perhaps when the warranty period has expired someone might have a look inside the unit and see if it can be done as on the ProTrack Scanner.
It would be possible but a bit destructive rerouting the connection between the main PCB and the Scanner PCB. I guess if you were not using a prober you could use the BNC connectors on the back as your external connections.