Quote from: fetron on January 10, 2023, 02:45:44 AM1. Warm up the Protrack for 30 minutes.this information is gold, thank you again.
2. Enter the Calibration menu.
Press Help button. Do not release it yet.
Press Menu button. Do not release it yet.
Press right side 4. button and release all three buttons.
Here, press the Auto on the right side to start the calibration.
The procedure runs for 20 minutes.
Run Workstation 3.5.4 with Win XP only.
Run the meter check in a Workstation (for this you need a 6.5-digit dmm, e.g. Agilent 34401)
Then try loop calc.
Maybe all values will be good.
, I adjusted it again. this time I went to a screen that had "store" function in there and pressed it to let unit store its setting. then turned it off / on.
same thing the trace was back, A Little Over To The Lefttttt.
, any help would be greatly appreciated. thank you.