Huntron 2000 110v to 220v and 50-60hz jumper

Huntron 2000 110v to 220v and 50-60hz jumper
« on: August 31, 2014, 08:43:40 am »
Hi!
I have a Huntron 2000 115v and I want to change it to 220v, I found the 115v - 220v internal switch but I can't find JP1 to change from 60hz to 50hz. Can you help me please?
« Last Edit: August 31, 2014, 09:45:05 am by mashak »

admin

  • Huntronian
  • 400
    • HUNTRON.COM
    • Email
Re: Huntron 2000 110v to 220v and 50-60hz jumper
« Reply #1 on: September 02, 2014, 07:35:34 am »
Welcome to the Support Forum Machak.

The Tracker 2000's with a serial number of less than 18000 did not have the 50/60 Hz jumper. On older 2000's the 50/60 Hz frequency uses the line frequency so you should be good. I suspect you have a newer manual but an older Tracker.
Thanks.

Admin
The most important troubleshooting tool is your eyes.

Re: Huntron 2000 110v to 220v and 50-60hz jumper
« Reply #2 on: September 02, 2014, 09:14:24 am »
Thank you very much!

Regards,
Christos